************************************************************************ MEMORANDUM June 29, 1995 DRAFT To: Mike Baker From: Don Cortes Re: Immunoassay vs. Mobile Laboratory I have had the opportunity to review materials regarding SW-846 Draft 4020, a rapid screening procedure for Polychlorinated Biphenyls (PCBs) using immunoassay. Draft 4020 is an inexpensive, relatively simple procedure for use in screening soil samples for polychlorinated biphenyls (PCBs) at a user's defined detection level. However, its use at Lemon Lane Landfill requires careful consideration of some potential problems. What follows is a comparison of using Draft 4020 on-site to another potential option, EPA Method 8080 using an on-site mobile laboratory. I wish to first point out that EPA Draft 4020 has not yet been promulgated, although it has been field tested and is expected to be included as a method in Revision 3 of SW-846 this year. EPA Method 8080 is an official method, and has been extensively employed since the early 1980s. This having been said, I will refer to the methods as simply EPA 4020 or EPA 8080 in the rest of this memorandum. Turn Around Time: EPA Draft 4020 appears to be a significantly faster procedure. Vendors indicate that 1 sample can be tested in 30 minutes, and as many as 17 samples in 60 minutes. I suspect 10 samples in 60 minutes is a more reasonable expectation, however. EPA Method 8080 can be performed at the rate of 1 sample in 60 minutes, or 20 samples in 24 hours, using a rapid extraction procedure that saves sample preparation time at the expense of an increased detection limit. The detection limit using this technique is 1 ppm in soil, which should be acceptable for most applications. As will be discussed further on, EPA Draft 4020 requires confirmation using EPA Method 8080 on at least 10% of the samples found to contain PCBs. Because this confirmation must be sent to an offsite laboratory, the effective turn around time may be the same for both methods, when considering turn around time in terms of a daily batch of samples. Of course, samples tested by EPA Method 8080 do not need confirmation. Cost EPA 4020 is significantly cheaper than EPA 8080. I estimate a total cost for this project of $15,000, which includes: 1. $2500 for accessories. This includes all the basic accessories needed, including a photometer. 2. $1820 for equipment to test 250 samples. 3. $2400 for analyst at $15/hour, 20 days. 4. $5000 for 25 confirmation samples at $200 per sample (rush charge). 5. $2500 for method validation. This includes the cost of 15 samples at $150/sample for EPA 8080, and an analyst at $15/hour. A mobile laboratory would cost about $35,000 for 20 days, 250 samples. This cost is based on my experience as a mobile laboratory manager, and consultation with current mobile laboratory contractors. Quality While EPA 4020 may offer the advantages of quick turn around time and low cost, it pales to EPA 8080 in terms of quality. 1. Accuracy EPA and Immunoassay vendors explicitly state that EPA 4020 results are to be considered estimates only. For accurate, precise results, EPA 8080 must be employed for each and every sample. This is due to the fact that: A. Quantitation is performed using a standard that is probably dissimilar to the actual composition of the PCB mix found. There is no way to choose a similar standard without knowing what type of PCB is present in the sample, and EPA 4020 does not give that information. B. There is no way to detect many conditions that lead to false positives or false negatives with EPA 4020. In this case, a false positive results when PCBs are reported present when actually they are not. This may result from any compound other than a PCB that has a binding affinity for the antibodies that make up the solid phase of the reaction chamber. A false positive may also result if any compounds or oils are present that prevent the enzyme conjugate from binding to the antibody sites. There is no way to "see" this happening in the reaction chamber. With EPA 8080, on the other hand, one can "see" the PCBs by looking at the chromatogram. Each Arochlor mixture displays a unique pattern that no other chemical or mixture can exactly reproduce. The chances of a reporting a false positive by EPA 8080 are therefore extremely low. A false negative is a result of "not detected" when in fact PCBs are present at greater than the user defined detection level. Erroneous results can be caused by any compounds which are present in the sample which interfere with the affinity that a PCB molecule has for an antibody site, or any compounds that restrict the availability of an antibody site for the target analyte. For example, the presence of oils can alter the statistical distribution of enzyme conjugate and PCBs that are competing for the antibodies. Since calculations assume a fair randomness in this competitive inhibition reaction, the presence of oils can cause unpredictable results. With EPA 8080 however, surrogate standards are employed with every analysis in order to verify recovery of target analytes. When these surrogates fall outside of an acceptance range, the analysis is suspect and is repeated until the cause of the failure can be determined. Thus the chances of reporting a false negative by EPA 8080 are extremely low. I believe that the risk of reporting a false negative when using EPA 4020 on landfill material is greater than when analyzing soil, and thus is much higher than indicated in the method. This is because landfill material often contains high concentration of chemicals that may cause interferences. Of course, the method must first be validated using soil that is characteristic of the sample that is to be analyzed. Since landfill soil is non-homogeneous, this presents a tricky problem. Suffice it to say that validation of the method must be extensive in it's coverage of all possible interferences. 2. Information As has been previously stated, EPA 4020 provides no information as far as which mixtures of PCBs are present. This information is important in choosing appropriate calibration standards for accurate analysis and determining the likelihood migration in water systems and soil. EPA 8080 does provide this information. In addition, if trichlorobenzenes or tetrachlorobenzenes are part of the dielectric fluid used in the capacitors (and they often are), they can be detected with a slight modification of EPA 8080 GC temperature program. 3. Detection. The lowest detection limit achievable by EPA 4020 is 1 ppm. However, the conceptual sampling plan for Lemon Lane indicates that the detection limit will be set at 400 ppm. I presume that this is because the contractor believes that only a presence/absence result can be obtained. If this were true, the effective detection limit would be 400 ppm. This would compare with EPA 8080 detection limit of 1 ppm (rapid extraction technique). I wish to point out, however, that Millipore's 4020 kit comes with 5 detection levels. 5 detection levels can also be used with the Ensys 4020 kit, and the additional cost would be no more than $50 for the whole project. There is absolutely no good reason not to use all 5 of these levels. A detection level could be set at 1, 50, 100, 400, and 1000 ppm, for example. Each sample would need to be read out a maximum of 2 extra times per sample, at the cost of an additional 30 seconds per sample. While exact quantitation would still not be achieved, the result could be reported as either: Less than 1 ppm Between 1 and 50 ppm Between 50 and 100 ppm Between 100 and 400 ppm Between 400 and 1000 ppm, or Greater than 1000 ppm . This is as opposed to the options of : Less than 400 ppm Greater than 400 ppm. General Quality Assurance Issues: It is very important that proper quality assurance protocol be followed. Some important points: 1. The test kits have not been demonstrated over 90¿F. A basic characteristic of antibodies and enzymes is that they denature (rendering them permanently useless) above a certain temperature. Precautions should be taken to insure that the temperature of the tubes and conjugate solution at no time exceed 90¿F prior to or during analysis. Thus, the equipment should not be kept in the car on a sunny day, in the field, or in a cooler that is in direct sunlight. In any case, the temperature should be monitored. 2. The accuracy and precision of this method must be assessed as part of the validation study. This must be performed under the same conditions as the analysis of the samples and at no more than 10 times the anticipated detection limit. Method Detection Limit (MDL) studies must take precision into account. 3. Each analyst must be qualified to perform the tests. There should be documentation as to how proficiency has been demonstrated. All the important points of a quality assurance program cannot be covered here. However, a comprehensive quality assurance program must be in place if EPA 4020 is to be employed, just as if testing were being performed in a mobile laboratory or an off-site laboratory. Conclusion: Because I do not know the objectives of the Lemon Lane sampling plan, I cannot comment on the appropriateness of either EPA 4020 or EPA 8080. It is clear, however, that EPA 4020 provides an inexpensive and quick method to determine the presence or absence of PCBs at user defined levels, at the cost of a slight to moderate reduction of confidence in analytical results. Table 1: Comparison of Immunoassay vs. Mobile Laboratory for PCBs in Soil Immunoassay Mobile Laboratory TURN AROUND TIME 1 samples/30 minutes 1 sample/ 60 minutes 17 samples/60 20 samples/24 hours minutes(?) COST $15,000 - TOTAL $35,000 - 20 days ($2550 - Accessories) ($1820 - 250 Samples) ($2400 - Analyst for 20 days) ($5000 - 25 Confirmation samples @ $200/sample) ($2500 - Method Validation) QUALITY Accuracy Estimate only High accuracy Detection 400 ppm 1 ppm (rapid extraction) Qualitative Identifies Arochlor PCB yes/no Risk of False High Extremely low Positives Risk of False Moderate Extremely low Negatives ***********************************************************************8